I am getting a strange mismatch between different equipment.
I had a low-Alk problem caused by a bad doser, but the Alk was not really
going back up like I expected. So I tried a different kit, and it did not match.
I thought it might be a simple reagent issue, so I tested both against
a fresh batch of instant ocean, and they matched there.
So now I am even more confused.
Hanna Electronic tester:
My tank = 7.4
Instant Ocean = 10.1
(I double checked ppm to dkh math)
Salifert test kit:
My tank = 8.3
Instant Ocean = 10.2
From experience and other places, that 10 dkh is expected on IO.
I am used to some difference between testers, but almost 1.0 is really huge.
I ran the test on my tank twice, being very careful the second time. Careful cleaning of equipment,
shaking reagents, the works. Values repeated within about .1.
Reagents for both kits are inside expiration date.
Ideas????
Which one should I believe??
My only theory is that it is some total alkalinity versus carbonate measurement issue.
The electronic versus titration do measure that differently.
I normally do not do many water changes, so perhaps Borates or something are accumulating.
I had a low-Alk problem caused by a bad doser, but the Alk was not really
going back up like I expected. So I tried a different kit, and it did not match.
I thought it might be a simple reagent issue, so I tested both against
a fresh batch of instant ocean, and they matched there.
So now I am even more confused.
Hanna Electronic tester:
My tank = 7.4
Instant Ocean = 10.1
(I double checked ppm to dkh math)
Salifert test kit:
My tank = 8.3
Instant Ocean = 10.2
From experience and other places, that 10 dkh is expected on IO.
I am used to some difference between testers, but almost 1.0 is really huge.
I ran the test on my tank twice, being very careful the second time. Careful cleaning of equipment,
shaking reagents, the works. Values repeated within about .1.
Reagents for both kits are inside expiration date.
Ideas????
Which one should I believe??
My only theory is that it is some total alkalinity versus carbonate measurement issue.
The electronic versus titration do measure that differently.
I normally do not do many water changes, so perhaps Borates or something are accumulating.